Hi, updated KStarts/Ekos/Indi yesterday from the stellermate repos and now my Optec TCS-S3 focuser operation is broken. It was maybe a year since my last update so that's unfortunate as it seems quite a few changes. Very basic operation is broken, it seems to operate in relative mode only, and in only a single direction . I've attached a log with just the focuser in the profile. Connected the focuser and tried to do an absolute move to 1000 from the driver control panel. It does a +1000 relative move to 6200 from its original position of 5200.
I fired on an old version on my mac, 2.9.6 vs 3.2.2 on ubuntu. It worked fine, enclosed is the log for that as well. The absolute/relative behavior seems to be the root of the problem.
David
Thank you for the report. Can you please try the INDI nightly later tonight and submit the logs again? The TCFS driver version should be v0.2 (you can verify that in the connection tab.
Seems inoperable now. Waited 15+ min for maybe a time out, didn't get one. In looking at the logs I don't understand the new calls and responses having to do with temperature and "Mode 1". I'm suppose to be in manual mode, just commanding position, nothing to do with temp. Log attached.
Ok looks like temperature readout failure affects this. So I've updated to version 0.3 to take care of it. Please try again later tonight or compile directly from Github.