Hi everyone
Sometimes the analyze tab displays the temperature of the camera sensor. Other times it displays ambient.
It seems to depend upon the sequence of events I use to start the cooler, but I can't find the magic sequence:(

The cooler is working, as seen on both the INDI control panel and the EKOS camera tab.
is there a foolproof way to do this so that the sensor temperature is always displayed?

Not a show stopper by any means. Just a mild inconvenience.

Cheers and clear skies,
Steve

logs:
drive.google.com/drive/folders/1W8R3XW85...onaxozgF?usp=sharing
Last edit: 10 months 1 week ago by alacant.
10 months 1 week ago #93538

Please Log in or Create an account to join the conversation.

  • Posts: 970
  • Thank you received: 94
Anyone?

Here's a session where analyze shows only ambient temerature. I've included the the verbose log too.
drive.google.com/drive/folders/12t16M0gM...NzNUU?usp=drive_link

Thanks,
Steve
8 months 1 week ago #94958

Please Log in or Create an account to join the conversation.

  • Posts: 1224
  • Thank you received: 566
Steve,

Analyze gets the temperature from the focus module.
The focus module has a UI switch for its temperature source, as shown below.

Hy

The following user(s) said Thank You: alacant